Quote:
clubfoot wrote:
Once you load the RS compensation file "uncheck" C Weighted SPL Meter as shown in Otto's pic. |
Are you sure about this?
The calibration files only have low frequency values upto about 80 hz.
The documentation says, when using a mic calibration file, REW will use the calibration values when provided and use the standard C-Weigting outside the calibration range.
If you turn off C-Weighting, won't you get inaccurate results above 80 Hz, especially above 8Hz?
Adri.