| SPL Meters | Mic's | Calibration Strange RS SPL meter calibration resultsDiscuss Strange RS SPL meter calibration results in the Subwoofer Equalization | Calibration forum; Strange RS SPL meter calibration results A friend of mine wanted to have his newer type analog Radio Shack SPL meter calibrated, so he sent it ... |
|
| | Thread Tools |
| | #1 (Link) | |||
| Strange RS SPL meter calibration results A friend of mine wanted to have his newer type analog Radio Shack SPL meter calibrated, so he sent it to me. I noticed some strange behaviour while calibrating it against to my professionally calibrated microphone. The amount of correction needed wasn't anywhere near to what Sonnie's correction file suggest for this type of meter. I believe a few graphs explain it better. All the measurements were done with the REW (4.00), I merely used the TrueRTA to show them. The first graph shows the difference between the correction files for my (i.e. my friend's) new analog meter and Sonnie's new analog meter. Quite a difference! ![]() The second graph shows the difference between the correction files for my new analog meter, my old analog meter, and Sonnie's old analog meter. As you can see, they all match reasonably well (within ~3 dB). My new analog meter seems to be only a little bit better than the old ones. ![]() So what does this all mean? Assuming both Sonnie and I took accurate measurements, some of the new analog meters are better then the others. The worse ones aren't much better than the old analog ones. This will definitely cause some strange measurement results when all the people having the new analog meter are using the same correction file. Their frequency response can look like having much more low frequency roll-off than it in reality has. So it looks like some of the new analog meters are still having the insides of the old analog meter. How do we separate them? I took a picture of the QC stamp, which reads "10A05". Could people post their QC codes for all the RS meters they have? Here's the list: New analog (soft-look) 04A06 (tatkinson) 12A05 (Chrisbee, toecheese) 10A05 (Ilkka's friend) 06A05 (WillyD, Average Joe) 4A05 (norpus) 02A04 (JRace) Old analog (boxy) 07A06 (Jeffeory) 11A05 (daxie) 12A03 (Ilkka) 10A02 (Peter De Smidt) 09A01 (snatcher) 9A4 (Chrisbee) ![]() | |||
|
| | |
| | |
| | #2 (Link) | ||||
| Re: Strange RS SPL meter calibration results Quote:
I seem to remember when Chrisbee said his new analog didn't give him anywhere near as good a response as he was measuring with the old meter. I wondered at the time why they weren't the same... This could be it. I do have to agree with something you've commented on before, and that's if you want to test down to 10hz it is probably a better idea to get a good microphone like many of us have resorted to.. brucek | ||||
|
| | #3 (Link) | |||||
| Re: Strange RS SPL meter calibration results Quote:
Quote:
![]() | |||||
|
| | #4 (Link) | |||
| Re: Strange RS SPL meter calibration results Recently purchased soft-look analogue: QC 12A05 TEC 12 year old (?) boxy shaped analogue: QC 9A4 TEC I shall do an REW sub response curve for each meter without loading any meter calibration files tomorrow. | |||
|
| | #5 (Link) | ||||
| Re: Strange RS SPL meter calibration results Quote:
I'd tape the two mics together..... brucek | ||||
|
| | #6 (Link) | |||
| | Re: Strange RS SPL meter calibration results This could definitely throw a kink in things, especially for those purchasing newer analog meters. I do have the $99 Galaxy CM-140 in hand, but haven't done any measurements with it. It may be that we simply need to advise people that they will need to either have someone check their RS meter for accuracy or purchase one of the CM-140's, if it turns out to be as accurate as we suspect it will be. | |||
|
| | #7 (Link) | |||
| Re: Strange RS SPL meter calibration results Okay. Nearfield measurements at 4" from the dustcap of bottom driver of my IB. Both meters carefully calibrated individually to 80dB prior to measurement. Meter calibration file in REW cleared prior to each measurement to ensure no cal file loaded. Here's the old analogue meter. ![]() Here's the new analogue meter. ![]() Like two peas in a pod!!! Last edited by Chrisbee; 01-16-07 at 03:48 AM. | |||
|
| | #8 (Link) | |||
| Re: Strange RS SPL meter calibration results Yep...... I guess you better use the cal file for the old meter. Ilkka has made an interesting discovery? What can we do about it? brucek | |||
|
| | #9 (Link) | |||
| Re: Strange RS SPL meter calibration results Those of us with two meters at least have a control (of sorts). The chances of two meters being so alike is er.... very unlikely. ![]() It certainly explains why my new meter behaves exactly like old meter when using the old RS analogue cal file. But can we trust Sonnie to have got the old RS analogue cal file right? How many were in his sample? Does this mean I can now have my bass back which Sonnie stole by giving us the new rs analogue file? ![]() ![]() | |||
|
| | #10 (Link) | ||||
| Re: Strange RS SPL meter calibration results Quote:
And might I say, that's a fine looking NF response. When you put the correction file in, you'll have loads of low frequency gain! How does it look at the LP? | ||||
|
| | #11 (Link) | |||||
| Re: Strange RS SPL meter calibration results Quote:
Quote:
![]() | |||||
|
| | #13 (Link) | ||||
| Re: Strange RS SPL meter calibration results Quote:
Sonnie got two new digital meters, two new analog meters and his own old analog to compare against the ECM8000. A new file was created for the old RS meter against his single old RS meter. A new file was created for the new analog meter against two new analog meters. A new file was created for the new digital meter against two new digital meters. The decision was made that if two meters compared the same, then that was as large a sample as we would use. The two new analogs tracked well and the two new digitals tracked well. Arguable a small sample, but since Sonnie paid for these himself, I think it was good enough. Once completed, Sonnie did a retest to compare the final results against his ECM. It was close enough and we posted the results for anyone who dared to use them..... Here is a jpg of a final test of the ECM8000 against the two new digital and two new analog with their posted cal files loaded. This pic didn't have the old analog in it, but it was just as accurate when compared with its cal file at the time. It would indeed appear though, that they may have switched the mic elements in the new ones at some time after they started producing them? What can you do. final_four_meters_with_new _cals_loaded.jpg brucek | ||||
|
| | #14 (Link) | ||||
| Re: Strange RS SPL meter calibration results Quote:
![]() Thanks for the explanation brucek. Last edited by Chrisbee; 01-16-07 at 08:41 AM. | ||||
|
| | #17 (Link) | |||||
| Re: Strange RS SPL meter calibration results Quote:
Quote:
Can you repost it with vertical scale 45dB - 105dB and horizontal 10Hz - 200Hz..... brucek | |||||
|
| | #18 (Link) | ||||
| Re: Strange RS SPL meter calibration results Quote:
I should have scaled to 800 not the 880 I have been using. I have also lifted the target curve to 80dB which is my calibration level in REW. Last edited by Chrisbee; 01-16-07 at 08:57 AM. | ||||
|
| | #21 (Link) | |||
| Re: Strange RS SPL meter calibration results We can't all be experts like you, Ilkka. ![]() I now have a whole mass of completely meaningless graphs because you questioned the old V new RS meter. I now have whole threads which are completely meaningless because they are based on the wrong meter cal files. ![]() But I bet my bass is better than yours even if I can't manage a simple graph. ![]() | |||
|